We are using the following scenario:

 

03 inspection type with inspection points 100

Sampling procedure has got valuation mode 700 - mean value
within tolerance limits

When we record results for 2 or 3 inspection points with some
of the results out of spec and then go to make UD, in the list of chars
relevant for UD the system summarizes recorded values. If a test result from
the last insp point is out of spec, then it is displayed in the list. If the
test result is in spec in the last insp point, then char is not listed, even
though there was a result out of spec recorded in one of the insp points.

For example, test A with lower limit 1 and upper limit 10

Insp point 1 test result = 9

Insp point 2 test result = 15

Insp point 3 test result = 9

 

In UD screen, test A is not listed as char relevant for UD,
even though for insp point 2 result is out of spec. And the average of all
three insp points is also out of spec.

What the business would like to see is that if a single
result is recorded out of spec for any of the insp points, then the char should
be displayed in chars relevant for UD with a red cross next to it.

Does anybody know what drives SAP logic to look at the last
insp point during UD? And can it be changed through config?

Inspection point identifier
100 does
not have Last Inspected Sample Is Valid

Thanks in advance

Accepted Solutions (1)

Accepted Solutions (1)

former_member42743
Active Contributor
0 Kudos

You should also review this thread very carefully.  I believe you might find it helpful.

http://www.sapfans.com/forums/viewtopic.php?f=27&t=360976

FF

former_member338769
Participant
0 Kudos

Thanks Fire Fighter.

I ran into this same post and followed the suggested solution by changing the FM in valuation mode used for inspection points - it worked a treat!

former_member42743
Active Contributor
0 Kudos

Yep.. I had forgotten about that post!  I swear I keep forgetting more and more stuff as time goes on.  Glad my alter ego worked on that post!

FF

Answers (1)

Answers (1)

former_member42743
Active Contributor
0 Kudos

In configuration, for the inspection point identifier at the bottom of the screen is a field:

"No. insp. points for char. summariz."

Make sure that has a number larger than the number of inspection points you expect.

Below that is a check box. "Last inspected sample is valid".  Make sure this is NOT checked.

I suggest you read the help file available for each of these fields.

Lastly, make sure each of your inspection points is valuated and closed.